PID Insulation Tester
View Large
Product Name : PID Insulation Tester
Product Origin : Hong Kong S.A.R., China
Product Category : Kikusui
More Products : Click here!
Contact Details : 運揚有限公司 WANCO LIMITED
Product Code : TOS7210S
Product Details

PID Insulation Tester



比例積分微分(控制器)


Model : TOS7210S

 

 

 

The Kikusui TOS7210S

 


* Specification

 


Output section

Output voltage range 50 V to 2000 V x 1 V

Accuracy ± (1.5 % of setting + 2 V)

Maximum rated output 2 W (2000 V/1 mA)

Maximum rated current 1 mA

Output terminals Output type Floating

Isolation voltage ± 1000 Vdc (The terminal that polarity is set to positive polarity)

+ 1000 Vdc and -3000 Vdc (The terminal that polarity is set to negative polarity)

Ripple 2000 V/under no load 20 Vp-p or less

Maximum rated load 20 Vp-p or less

Voltage regulation 1 % or less (Maximum rated load → No load)

Short-circuiting current 2 mA or less (Instant 200 mA or less)

Output rise time 60 ms or less (10 % to 90 %, no load)

Discharge function Forced discharge at the end of test (discharge resistance: 20 kΩ)

 

Voltmeter

Measurement range 0 V to 2400 V x 1 V

Accuracy ± (1 % of reading +1 V)

 

Resistance meter

Measurement range 0.01 MΩ to 5000 MΩ (In the range of over 100 nA to a maximum rated current of 1 mA)

Display □ . □□ MΩ [R < 10.0 MΩ] □□ . □ MΩ [10.0 MΩ ≤ R < 100.0 MΩ]

□□□ MΩ [100.0 MΩ ≤ R < 1000 MΩ] □□□□ MΩ [1000 MΩ ≤ R ≤ 5000 MΩ]

(R = measured insulation resistance)

Accuracy *1 ± (10 % of reading) [100 nA < i ≤ 200 nA]

± (5 % of reading) [200nA < i ≤ 1 μA]

±(2 % of reading) [1 μA < i ≤ 1 mA]

( i= measured output-voltage value/measured resistance value)

Measurement range The current measurement range is selectable between AUTO and FIX.

AUTO Automatically changes the current measurement range

according to the measured current value.

FIX Fixes the current measurement range based on the output voltage set value

and LOWER set value (in UPPER OFF status).

Holding function Holds the resistance value obtained at the end of testing while a PASS judgment is being output.

 

Ammeter

Measurement range 0.000 μA to 1900 μA

Display □ . □□□ μA [i < 10.00 μA] □□ . □□ μA [10.00 μA ≤ i < 100.0 μA]

□□□ . □ μA [100.0 μA ≤ i < 1000 μA] □□□□ μA [1000 μA ≤ i]

( i= measured current value)

Accuracy *2 ±(4% of reading +0.005 μA) [i < 10.00 μA]

±(4% of reading +0.005 μA) [10.00 μA ≤ i < 100.0 μA]

±(2% of reading +0.005 μA) [100.0 μA ≤ i < 1000 μA]

±(2% of reading) [1000 μA ≤ i]

( i= measured current value)

Measurement range The current measurement range is selectable between AUTO and FIX.

AUTO Automatically changes the current measurement range according to the measured current value.

FIX Fixes the current measurement range based on the output voltage set value

and LOWER set value (in UPPER OFF status).

 

Judgment function

Judgement

method/action

UPPER FAIL Judgement If a resistance value equal or less than the lower resistance is detected,

the tester shuts off the output and returns an UPPER FAIL judgment.

LOWER FAIL Judgement

If a resistance value equal or less than the lower resistance is detected,

the tester shuts off the output and returns a LOWER FAIL judgment.

Note that no judgment is made within the judgment wait time (WAIT TIME) after the start of the test.

 

Time

Setting range for the test duration 0.5 s to 999 s (Consecutive operation by setting TEST TIME as OFF)

Setting range for the wait time 0.3 s to 10 s (TEST TIME > WAIT TIME)

Accuracy ±(100 ppm + 20 ms)

 

SIGNAL I/O D-SUB 25-pin connector on the rear panel

Input

specifications

High-level input voltage 11 V to 15 V

All input signals are active Low controlled.

The input terminal is pulled up to +12 V using a resistor.

Opening the input terminal is equivalent to inputting a high-level signal.

Low-level input voltage 0 V to 4 V

Low-level input current -5 mA maximum

Input time width 5 ms minimum

Output

specifications

Output method Open collector output (4.5 Vdc to 30 Vdc)

Output withstand voltage 30 Vdc

Output saturation voltage Approx. 1.1 V (at 25°C)

Maximum output current 400 mA (TOTAL)

 

ANALOG OUT Outputs the measured resistance, measured current and voltage, and current range in DC voltage.

Measured resistance V0 = log

Rx: (1 MΩ: 0.3 V, 10 MΩ: 1.04 V, 100 MΩ: 2.00 V, 1000 MΩ: 3.00 V, 10000 MΩ or more: 4.00 V)

Output impedance: 1 kΩ

Measured current Renge1: V0 [V]= measured value [μA]/ 512 Renge3: V0 [V]= measured value [μA]/8

Renge2: V0 [V]= measured value [μA]/64 Renge4: V0 [V]= measured value [μA]

COM Analog output-circuit common

Accuracy ±(2 % of full scale)

RS232C D-SUB 9-pin connector on the rear panel (compliant with EIA-232-D)

All functions other than the POWER switch and KEY-LOCK

Baud rate 9600 bps/19200 bps/38400 bps (data: 8 bits; parity: none; stop bit: 2 bits fixed)

REMOTE 6-pin mini-DIN connector on the front panel

The optional remote controller RC01-TOS or RC02-TOS is connected to control remotely

starting/stopping of a test (note that a DIN-mini DIN adapter is required).

Display 7-segment LED, 4-digit voltage display, 4-digit insulation resistance display,

4-digit current display, and 3-digit time display

Memory function A maximum of 10 types of test conditions can be stored in memory

TEST MODE MOMENTARY A test is conducted only when the START switch is pressed.

FAIL MODE Disables cancellation of FAIL judgment using a stop signal via remote control.

DOUBLE ACTION Starts a test only when the STOP switch is pressed and the START switch is

pressed within approximately a half-second.

PASS HOLD Allows the time of holding PASS judgment to be set to 0.2 s or HOLD

KEYLOCK Places the tester in a state in which no keystroke other than the START/STOP switch is accepted

 

Environment

Installation location Indoors and at altitudes up to 2000 m

Warranty range Temperature/Humidity 15 °C to 30 °C/20 % rh to 80 % rh (no condensation) (59 °F to 86 °F )

Operating range Temperature/Humidity 0 °C to 40 °C/20 % rh to 80 % rh (no condensation) (32 °F to 104 °F )

Storage range Temperature/Humidity -20 °C to 70 °C/90 % rh or less (no condensation) (-4 °F to 158 °F )

 

Power requirements

Nominal voltage range(allowable voltage range) 100 Vac to 240 Vac (85 Vac to 250 Vac)

Power consumption At rated load 30 VA maximum

Allowable frequency range 47 Hz to 63 Hz

Insulation resistance 30 MΩ or more (500 Vdc) (AC LINE to chassis)

Hipot 1500 Vac for 1 second, 10 mA or less (AC LINE to chassis), 3000 V AC for 1 second (A, B terminals to chassis)

Ground bond 25 Aac/0.1 Ω or less

Dimensions (mm (inch)) (maximum dimensions)

/Weight

214 (8.43") W × 81(3.19") (115 (4.53")) H × 340 (13.39") (385 (15.16")) D

/Approx. 2 kg (Approx.4.41 lbs)