Maximum rating
voltage
AC 5.0 kV
DC 6.0 kV
Number of channels 4 (Each channel is settable to HIGH, LOW, or OPEN.)
Maximum number of scanners connected 4 scanners,Channel numbers are determined in order of connection to the TOS9200/9201 tester.
1 st scanner CH1 to CH4 2 nd scanner CH5 to CH8 3 rd scanner CH9 to CH12 4 th scanner CH13 to CH16
Contact check function None
Lamps and LEDs POWER Lights as it is interlocked with the POWER switch of the TOS9200/9201 tester
DANGER Lights as it is interlocked with the DANGER lamp of the TOS9200/9201 tester
CHANNEL Lights during a test at each channel HIGH: red; LOW: green; Under contact check: orange
Power requirements
Nominal voltage range (allowable voltage range) 100 V to 120 V AC/200 V to 240 V AC (85 V to 132 V AC/170 V to 250 V AC) Automatic switching
Power consumption In READY state Approx. 12 VA
During test 40 VA maximum
Allowable frequency range 47 Hz to 63 Hz
Insulation resistance 30 M Ω or more (500 V DC) [between the AC LINE and chassis]
Hipot 1390 V AC, 2 seconds, 10 mA or less [between the AC LINE and chassis]
Ground bond 25 A AC/0.1 Ω or less
Electromagnetic compatibility (EMC) (*2) Conforms to the requirements of the following directive and standard.
EMC Directive 2014/30/EU, EN61326-1, EN61000-3-2, EN61000-3-3
Under following conditions
1. Used test leadwire TL07-TOS which is supplied. 2. No discharge occurs at outside of the tester.
3. Used the shielded cable which length is less than three meters when the SIGNAL I/O is used.
Safety (*2,3) Conforms to the requirements of the following directive and standard.
Low Voltage Directive 2014/35/EU, EN61010-1, Class I, Pollution degree 2
Environment
Installation location Indoors and at altitudes up to 2000 m
Warranty range Temperature 5 °C to 35 °C
Humidity 20 %rh to 80 %rh (no condensation)
Operating range Temperature 0 °C to 40 °C
Humidity 20 %rh to 80 %rh (no condensation)
Storage range Temperature -20 °C to 70 °C
Humidity 90 %rh or less (no condensation)
Dimensions 430[16.93 inch](435[17.13 inch])W × 88[3.46 inch](105[4.13 inch])H × 370[14.57 inch](415[16.34 inch]) Dmm
Weight Approx. 6.5 kg (Approx.14.33 lbs)
Accessories
AC power cable 1 pc.
High-voltage test leadwires, red 4 pc. (1.5 m each)
High-voltage leads for parallel connection 1 set (0.5 m each)
Interface cable 1 pc.(0.5 m)
Channel-indication stickers For the panel face: 1 sheet; for the test leadwires: 1
“HIGH VOLTAGE, DANGER” stickers 2 sheets
Fuses 2 pc. (including a spare contained in the fuse holder)
Operation Manual 1 copy