* Specification
Output block
Current setting range 3.0 Aac to 31.0 Aac x 0.1 A
(With respect to resistance resulting in output power of the maximum rated Output or less and an output terminal voltage of 5.4 V or less)
Accuracy ± (1% of setting + 0.2 A)
Maximum rated output 150 VA (at the output terminals)
Distortion factor 2% or less (with respect to 0.1 Ω pure resistance
load of 10 A or greater)
Frequency 50/60 Hz, sine wave (selectable)
Accuracy ±200 ppm
Open terminal voltage 6 Vrms or less
Output method PWM switching method
Output ammeter
Measurement range 0.0 Aac to 33.0 Aac x 0.1 A
Accuracy ± (1% of reading + 0.2 A)
Response Mean value response/rms value display (response time: 200 ms)
Holding function The current measured at the end of test is held
during the PASS or FAIL inteval
Output voltmeter
Measurement range 0.00 Vac to 6.00 Vac x 0.01 V
Accuracy ± (1% of reading + 0.02 V)
Response Mean value response/rms value display (response time: 200 ms)
Holding function The voltage measured at the end of test is held
during the PASS or FAIL inteval
Ohmmeter
Measurement range 0.001 Ω to 1.200 Ω
Resolution 0.001 Ω
Offset cancel function 0.000 Ω to 1.200 Ω (Offset ON/OFF function provided)
Accuracy ± (2% of reading + 0.003 Ω)
Holding function The resistance measured at the end of test is held
during the PASS interval
Pass/fail judgement function
Resistance value-based judgement Window comparator system
•If a resistance value equal to or greater than the upper
reference value is detected,a FAILdetermination is returned.
•If a resistance value equal to or less than the lower
reference value is detected, a FAIL determination is returned.
•If a resistance value has been judged as FAIL, the tester
shuts off the output and generates a FAIL signal.
•If the set time elapses without abnormalities, the tester
shuts off the output and generates a PASS signal.
Setting range for the upper rerence value (UPPER)
0.001 Ω to 1.200 Ω
Setting range for the upper rerence
value (LOWER)
0.001 Ω to 1.200 Ω x 0.001 Ω
Judgement accuracy ± (2% of UPPER + 0.003 Ω)
Calibration Calibration is performed with the rms value of the sine
wave, using a pure resistance load.
LED
PASS Lights for approximately 0.2 sec when the measured
value has been judged as PASS.It is lit continuously
when the PASS holding time is set to HOLD.
UPPER FAIL Lights if a resistance value equal to or greater than
the upper reference value is detected and judged FAIL.
LOWER FAIL Lights if a resistance value equal to or greater than the
upper reference value is detected and judged FAIL.
Buzzer •The buzzer sounds for the pass holding time has been
set if the measured value has been judged as PASS.
•The buzzer sounds continuously under the following
condition:
The measured value has been judged as PASS when the
PASS holding time is set to HOLD.
The measured value has been judged as UPPER FAIL.
The measured value has been judged as LOWER FAIL.
•The buzzer volume for FAIL or PASS judgment are
adjustable.
Note that it cannot be adjusted individually since setting
is shared with the setting for PASS.
Time
TimeTest
Setting range 0.3 s to 999 s Timer ON/OFF function is available.
Accuracy ± (100ppm of setting + 20ms)
Environment
Operating environment Indoor use, Overvoltage Category II
Warranty range Temperature : 5°C to 35°C
Humidity : 20 %rh to 80 %rh (non condensing)
Operating range Temperature : 0°C to 40°C
Humidity : 20 %rh to 80 %rh (non condensing)
Storage range Temperature : -20°C to 70°C
Humidity : 90 %rh or less (non condensing)
Altitude Up to 2000 m
Power requirement
Allowable voltage range 85 Vac to 250 Vac
Power
consumption
At no load (READY) 60 VA or less
At rated load 280 VA max.
Allowable frequency range 47 Hz to 63 Hz
Insulation resistance 30 MΩ min. (500 Vdc), between AC line and chassis
Withstanding voltage 1390 Vac (2 seconds), between AC line and chassis
Earth continuity 25 Aac/ 0.1 Ω max.
Safety
Conforms to the requirements of the following directive and standard.
Low Voltage Directive 2014/35/EU, EN 61010-1 (Class I, Pollution degree 2)
Electromagnetic compatibility (EMC)
Conforms to the requirements of the following directive and standard.
EMC Directive 2014/30/EU, EN 61326-1 (Class A), EN 55011 (Class A, Group 1),
EN 61000-3-2, EN 61000-3-3
Under following conditions
1. Used test leadwire (TL11-TOS for TOS6200A, TL12-TOS for TOS6210)
which is supplied.
2. Used the shielded cable which length is less than three meters when the SIGNAL I/O is used.
Physical dimensions (maximum) 430[16.93 inch] (455[17.91 inch]) W ×
88[3.46 inch] (140[5.51 inch]) H × 270[10.63 inch] (345[13.58 inch]) D mm
Weight Approx. 9 kg (Approx.19.84 lbs)
Accessories
AC power cord 1 piece
Test leadwire TL11-TOS 1 set
Short bar 2 pieces (These are inserted between the OUTPUT
and SAMPLING terminals.)
AC power fuse 2 pieces (2, including one spare in the fuse holder)
Operation manual 1 copy