Hipot and Insulation Resistance Tester
耐高溫和絕緣電阻測試儀
Model : TOS5302
The Kikusui TOS5302
* Specification
AC output section
Output range 0.05 kV to 5.00 kV
Accuracy ±(2 % of set + 20 V) when no load is connected
Setting range 0.00 kV to 5.50 kV
Resolution 10 V steps
Max. rated output *1 500 VA (5 kV/100 mA)
Max. rated voltage 5 kV
Max. rated current 100 mA (when the output voltage is 0.5 kV or greater)
Transformer rating 500 VA
Output voltage waveform *2 Sine
Distortion If the output voltage is 0.5 kV or more: 3 % or less (when no load or a pure resistive load is connected).
Frequency 50 Hz or 60 Hz
Accuracy ±0.5 % (excluding during voltage rise time)
Voltage regulation 10 % or less (when changing from maximum rated load to no load)
Input voltage variation ±0.3 % (5 kV when no load is connected; power supply voltage: 90 V to 250 V)
Short-circuit current 200 mA or more (when the output voltage is 1.0 kV or greater)
Output method PWM switching
DC output section
Output range
0.05 kV to 6.00 kV
Accuracy ± (2 % of set + 20 V)
Setting range 0.00 kV to 6.20 kV x 10 V STEP
Max. rated output 50 W (5 kV / 10 mA)
Max. rated voltage 6 kV
Max. rated current 10 mA
Ripple(TYP)
5 kV when no load is connected 50 Vp-p
Max. rated load 100 Vp-p
Voltage regulation 3% or less (When changing from maximum
rated load to no load)
Short-circuit current (TYP) 40 mA (when generation 6 kV output)
Discharge feature Forced discharge after test completion
(discharge resistance: 125 kΩ)
Start Voltage The voltage at the start of withstanding voltage tests can be set to 50% of the test voltage.
Limit Voltage The test voltage upper limit can be set . AC: 0.00 kV to 5.50 kV, DC: 0.00 kV to 6.20 kV
Output voltage monitor feature If output voltage exceeds the specified value + 350 V or is lower than the specified value - 350 V,
output is turned off, and protective features are activated.
Voltmeter
Analog
Scale 6 kV AC / DC f.s
Indication Average value response/rms scale
Digital
Measurement range 0.000 kV to 6.500 kV AC / DC
Display is kV
Accuracy V < 500 V: ±(1.5 % of reading + 20 V); V ≥ 500 V: ±1.5 % of reading
Response True rms, Average value response / rms display switchable
Hold feature After a test is finished, the measured voltage is retained until the PASS or FAIL judgment is cleared.
Ammeter Digital
Measurement range AC: 0.00 mA to 110 mA
Display 6.500 kV
i = measured current
i < 1 mA, 1 mA ≤ i < 10 mA, 10 mA ≤ i < 100 mA, 100 mA ≤ i
Accuracy 1.00 mA ≤ i: ±(1.5 % of rdng); i < 1.00 mA: ±(1.5 % of reading + 30 μA)
Response True rms, Average value response / rms display switchable
Hold feature After a test is finished, the measured voltage is retained until the PASS judgment is cleared.