Electrical supply range
230 V a.c. 230 V a.c. ± 10%
110 V a.c. 110 V a.c. ± 10%
Supply measurement
Voltage: ± 2% ± 1 V
Frequency: ± 1% ± 0.1Hz
Bond test
Open circuit voltage: 9 V a.c. ± 10% ± 0.1 V (Supply: 230 V 50Hz)
10 A Bond test current:
10 A rms ±20% ± 0.5A into 0.1 Ω
25 A Bond test current:
25 A rms ± 5% ± 0.5 A into 0.1 Ω
Earth bond resistance accuracy:
±5% ±3 digits (0 to 0.49 Ω)
±5% ±5 digits 0.5 to .99 Ω)
Earth bond resistance resolution:
10 mΩ (0 to 1.99 Ω)
Display range: 0 to 1.99 Ω
Bond test nulling: Up to 1.00 Ω
Adjustable test duration:
User selectable from 1 sec to 20 sec
Continuity test
Note:
The continuity test is a DC test performed automatically
in both positive and negative directions. The average of
the two results is shown.
Continuity test voltage:
4.0 V d.c. -0% +10% (open circuit)
Continuity test current:
200 mA -0% +10% ±5 mA (into 2 Ω load)
Continuity resistance accuracy:
±5% ±3 digits (0 to 0.49 Ω)
±5% ±5 digits 0.5 to .99 Ω)
Continuity resistance resolution:
10 mΩ (1 to 19.99 Ω)
Display range: 0 to 19.99 Ω
Continuity test nulling:
up to 9.99 Ω
Test duration: User selectable from 1 sec to 20 sec
Insulation test
Insulation test voltage:
250 V d.c. –0%/+25% open circuit
500 V d.c. –0%/+25% open circuit
(500 V d.c. across 0.5MΩ)
Insulation resistance accuracy (230 V):
±2% ±5 digits (0 to 19.99 MΩ)
±5% ±10 digits (20 to 99.99 MΩ)
Insulation resistance accuracy (110 V):
±5% ±5 digits (0 to 10 MΩ)
Insulation resistance resolution:
0.01 MΩ (0.10 to 99.99 MΩ)
Display range: 0 to 99.99 MΩ
Test duration: User selectable from 1 sec to 1 minute
Substitute leakage test
Test voltage: 40 V a.c. ±10%
Test frequency: Nominal mains 50/60Hz
Leakage current accuracy: ±5% ±5 digits
Leakage current resolution: 0.01 mA
Display range: 0 to 19.99 mA
Test duration: User selectable from 1 sec to 1 minute
Reading corrected to 230 V + 10% or 110 V + 10% a.c.
Differential leakage current
Test voltage and frequency: Nominal mains supply
Differential leakage current accuracy: ±5% ± 0.05mA (1.0mA
to 19.99mA)
±5% ± 0.1mA (0.1mA
to 0.99mA)
Differential leakage current resolution: 0.01 mA
Display/measurement range: 0.10mA to 19.99 mA
Touch current test
Test voltage: Nominal mains 115/230 V a.c.
Test frequency: Nominal mains 50/60Hz
Touch current accuracy: ±5% ±5 digits
Touch current resolution: 0.01 mA
Display range: 0 to 10 mA
Test duration: User selectable from 1 sec to 5 sec
Reading corrected to 230 V + 10% or 110 V + 10% a.c.
Operational test
Test voltage: Nominal mains 110/230 V a.c.
Accuracy: ±5% ±10 digits (0 to 99 VA)
±5% ±50 digits (100 VA - 999 VA)
±5% ±100 digits (1000 VA - 3700 VA)
Resolution: 1 VA (0 to 3700 VA)
Display range: 0 to 3.99 KVA
Reading corrected to 230 V or 110 V a.c.
Results show load VA
Extension lead test
Test includes insulation and bond tests.
Polarity test voltage: 12 V
Polarity: Lead OK
Live neutral S/C
Live neutral reversed
Live/neutral O/C
Flash test
Flash test voltage: 1500 V a.c nominal for Class 1
3000 V a.c nominal for Class 2
Flash test current: < 3.5 mA short circuit @ 253 V primary supply
voltage
Flash test breakdown current accuracy: ±5% ±5 digits
Flash test breakdown current resolution: 0.01 mA
Display range: 0 to 3.0 mA
Test duration: User controlled
Portable RCD test
RCD test voltage: Nominal mains 110/230 V
RCD test frequency: 50Hz
Test current accuracy: -8% to –2% (1⁄2 x I)
+2% to +8% (1 x I, 5 x I)
Trip time accuracy: ±1% ±5 digits
Trip time resolution: 0.1 ms
Display range: 0 to 1999 ms (1⁄2 x I)
0 to 300 ms (1 x I)
0 to 40 ms (5 x I)
Fuse test
Test voltage: 3.3 V
Warning: Audible beep if fuse is OK
Safety
PAT300 series meet the requirements of IEC 61010-1: 2001
Test leads meet the requirements of IEC 61010-031: 2002
Creepage and clearances for 300 volts to Earth Category II
Operating temperature 10 to +50 deg C
Storage Temperature: -20 to +60 deg C
Humidity 90%RH @ -10 to +30 deg C
75%RH @ +30 to +50 deg C
Weight
PAT350 (instrument only) 5.0 kg
Dimesions
Instrument only: 175mm (H) x 320mm (D) x 250mm (W)
With case 190mm (H) x 400mm (D) x 290mm (W)